Van Lysel MS.
Departments of Medicine,, University of Wisconsin, H6/333 Clinical Science Center, 600 N Highland Ave, Madison, WI 53792, USA. vanlysel@facstaff.wisc.edu
Radiographics. 2000 Nov-Dec;20(6):1769-86
In fluoroscopic/fluorographic systems, an image intensifier is optically coupled to recording cameras. The optical distributor is responsible for transmitting a focused image from the output phosphor of the image intensifier to the focal planes of the cameras. Each camera has an aperture, which is used to control the level of light reaching its focal plane. The aperture setting determines the patient x-ray exposure level and the image noise level. Increasing the x-ray exposure reduces image noise; reducing the x-ray exposure increases image noise. Fluoroscopic/fluorographic systems always include a video camera. The functions of the video system are to provide for multiple observers and to facilitate image recording. The camera head contains an image sensor, which converts the light image from the image intensifier into a voltage signal. The device used to generate the video signal is a pickup tube or a charge-coupled device sensor. The method used is raster scanning, of which there are two types: progressive and interlaced. The vertical resolution of the system is primarily determined by the number of scan lines; the horizontal resolution is primarily determined by the bandwidth. Frame rate reduction can be a powerful tool for exposure reduction.