Taubel JP, Schueler BA, Vrieze TJ, Gray JE.
Department of Diagnostic Radiology, Mayo Clinic, 200 First St SW, Rochester, MN 55905, USA. taubel.jerome@mayo.edu
Radiographics. 2001 Mar-Apr;21(2):507-18
A noninvasive method was developed for quantifying the overall contrast of fluoroscopic imaging systems within the clinical setting by using a simple phantom and common video test equipment. In this method, an acrylic phantom with four holes filled with varying amounts of air and aluminum is placed on the entrance exposure side of a patient-equivalent acrylic phantom. The air- and aluminum-filled holes provide a stepped gray-scale pattern that is displayed on the examination room viewing monitor when the phantom is fluoroscopically imaged under automatic brightness control. A video waveform monitor or oscilloscope is then used to quantify those video signal voltage levels as a contrast index value, which is defined as the maximum range of the video signal voltage levels of the gray-scale steps. The method is repeatable and allows quantification of the contrast of the imaging system. It can also be used to optimize video parameters, provide comparative data for quality control monitoring, and characterize overall contrast differences between systems. Experience with this method suggests that there is excellent correlation between the clinical perception of image contrast and the contrast index, with contrast index changes of approximately 15% being seen clinically.
Posted via PubMed for educational and discussion purposes only.
Link to PubMed Reference