Mount CJ, Gray JE.
Department of Diagnostic Radiology, Mayo Clinic, Rochester, MN 55905.

Radiographics. 1990 Nov;10(6):1049-54

Good screen-film contact is essential to obtain high definition in radiographic images. We compared the conventional test tool for measuring screen-film contact, which uses #8 mesh (eight wires per inch [about three wires per centimeter]), with a prototype tool of #40 copper mesh. The conventional tool did not adequately test screen-film contact for mammography, in which significantly higher resolution is obtained. We believe #40 copper mesh should be used to test screen-film contact of mammographic cassettes. We also found that entrapped air may reduce screen-film contact and, therefore, recommend that cassettes be loaded at least 15 minutes before they are exposed.