Miyati T, Kasai H, Shundo H, Imazawa M, Banno T, Ohba S, Itikawa K, Hirose Y.
Department of Central Radiology, Nagoya City University Hospital, Aichi, Japan.
Radiographics. 1996 May;16(3):595-602
A method was developed for performing T1-weighted magnetic resonance imaging with the rapid acquisition with relaxation enhancement (RARE) sequence by altering the excitation flip angle. This method was called variable flip angle turbo spin-echo (VF-TSE) imaging. When the effective echo time corresponds to the first echo, the resolution worsens as the echo train length becomes longer. For this reason, the echo train length was set at three, the repetition time (TR) was shortened (100- 200 msec) to decrease imaging time, and the initial flip angle was adjusted (120 degrees-140 degrees) to improve image quality. Another advantage of this method is that the initial flip angle can be reduced to below 90 degrees when a longer TR is needed. Measured signal intensities for VF-TSE imaging matched theoretic predictions. VF-TSE imaging yielded high contrast-to-noise and signal-to-noise ratios without sacrificing resolution. The VF-TSE technique was useful for breath-hold, three-dimensional, and cardiac synchronization imaging.